How Should A Meter For Tds Be Calibrated If The Dissolved Solids Are Not The Same As Those In The Calibration Solution?

Answer»

Making your own standard will yield the most accurate results. This is done by formulating a mixture of salts in relative proportions to those that simulate the solution being tested, then DISSOLVING the mixture into distilled water.

This should be done according to the FOLLOWING formula:

1 mg salt mixture/litre of distilled water = 1 ppm TDS,

(or)

X ppm TDS = X mg of salts + one litre of distilled water.

CHOOSE a ppm value for a calibration solution which is as close as possible to the expected ppm values of the test solutions. If the ppm content of the test solution is expected to vary a great deal, it is best to choose a ppm value for the calibration solution in the upper 1/3 of the expected TDS MEASUREMENT range.

  • Above mentioned Conductivity Analyzer Parameters may change from VENDOR to vendor

Making your own standard will yield the most accurate results. This is done by formulating a mixture of salts in relative proportions to those that simulate the solution being tested, then dissolving the mixture into distilled water.

This should be done according to the following formula:

1 mg salt mixture/litre of distilled water = 1 ppm TDS,

(or)

X ppm TDS = X mg of salts + one litre of distilled water.

Choose a ppm value for a calibration solution which is as close as possible to the expected ppm values of the test solutions. If the ppm content of the test solution is expected to vary a great deal, it is best to choose a ppm value for the calibration solution in the upper 1/3 of the expected TDS measurement range.