1.

Factors that reduce the dose to the patient include: a. Increased filtration b. Reduced collimation c. Using a grid with paediatric patients d. Using an intensifying screen e. Increasing the X-ray focus-to-patient distance

Answer»

a. True. 

b. False. Better collimation prevents unnecessary exposure to tissues and also acts to reduce scatter. 

c. False. As younger patients are smaller, a lower kV may be used and so scatter will be less and grids can be avoided. Because grids reduce the intensity of the main radiation beam, a grid will result in an increased dose to the patient by increasing the required mA. 

d. True. 

e. True. However, this can lead to a reduction in magnification.