Regarding CT: a. In scanograms, the scatter is high and spatial resolution is poor b. CT fluoroscopy is a low kV technique in which the table does not move c. In CT fluoroscopy, both the effective dose and the skin dose are less than in normal CT d. A scanogram is another name for a scout view e. CT fluoroscopy is used for biopsy needle placement |
|
Answer» a. False. In scanograms (topograms), the scatter is minimal and spatial resolution is poor. b. True. In CT fluoroscopy, the gantry rotates 180° without the table moving, and it is a low kV technique. c. False. In CT fluoroscopy the effective dose is less but the skin dose is higher because it is confined to a narrow region. d. True. A scanogram is another name for a scout view. e. True. |
|