1.

Regarding electrical safety and quality assurance (QA) in ultrasound: a. Electrical safety tests of the ultrasound unit should be performed annually b. QA of the ultrasound unit should be performed semi-annuallyc. The resolution in B-mode imaging may be tested with a string phantom d. Phantoms used for B-mode imaging are usually filled with distilled water e. The power output of the transducer may be measured with a radiation force balance

Answer»

a. True. 

b. False. The Royal College of Radiologists recommends in its ‘Standards for Ultrasound Equipment’ that testing should be performed annually or biannually. 

c. False. Resolution is usually tested using a test object containing a set of filament targets in a tissue-mimicking material phantom. A string phantom is used for testing the velocity in a Doppler mode. 

d. False. They need to be filled with a material in which the speed of sound (1540 m s-1 ), attenuation (0.5 dB cm-1 MHz-1 ) and density (1050 kg m-3 ) are close to that in a soft tissue.

e. True. The ultrasound emitted from a transducer exerts a force proportional to the output power, which can be ‘weighed’ on a special balance. A power of 1 W corresponds to approximately 68 mg.