1.

Which of the following is used as detector in improved infrared analyzer used for gas analysis?(a) PbSe(b) MgCl2(c) ZnCl(d) CuCl2The question was posed to me during an online exam.This intriguing question originated from Infrared Gas Analysers in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct answer is (a) PbSe

For EXPLANATION I WOULD say: PbSe is USED as detector in improved infrared ANALYZER used for gas ANALYSIS. It is a solid state detector.

2.

In analyzers based on gas density, if the sample carries a gas having a higher density than the detector which of the following occurs?(a) It will cause a net downward flow(b) It will cause a net upward flow(c) It splits into two beams(d) There will be no difference in flowI got this question in a national level competition.Question is taken from Analysers Based on Gas Density in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right CHOICE is (a) It will cause a net downward FLOW

The best I can explain: In analyzers based on gas density, if the SAMPLE CARRIES a gas having a higher density than the detector it will cause a net downward flow. It will obstruct flow in the lower path.

3.

In Paramagnetic oxygen analyzer, when no oxygen is present the magnetic force balances torque of the fibre.(a) True(b) FalseThe question was posed to me during an interview.My question is based upon Paramagnetic Oxygen Analyser topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct choice is (a) True

The BEST explanation: In Paramagnetic oxygen analyzer, when no oxygen is present the magnetic force balances TORQUE of the fibre. If oxygen is present, it WOULD DISPLACE the dumb-bell sphere.

4.

Magnetic wind analyzers are also known as which of the following?(a) Thermo-magnetic analyzers(b) Electro-magnetic analyzers(c) Thermo-analytical analyzers(d) Hot wire analyzersThis question was addressed to me in an interview for job.This interesting question is from Magnetic Wind Instruments in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right CHOICE is (a) Thermo-magnetic analyzers

Explanation: Magnetic wind analyzers are also KNOWN as Thermo-magnetic analyzers. In Europe, it is known as Magnetic wind analyzers. In the UNITED STATES, it is known as Thermo-magnetic analyzers.

5.

Which of the following occurs in Electrochemical cells used for the detection of hydrogen sulphide?(a) Change in resistance(b) Redox reaction(c) Oxidation-reduction reaction(d) Change in colourI have been asked this question during an interview.Query is from H2S Analysers topic in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right answer is (c) Oxidation-reduction reaction

The explanation: In ELECTROCHEMICAL cells, an oxidation-reduction reaction OCCURS in the presence of HYDROGEN sulphide. A current starts FLOWING and it is proportional to the CONCENTRATION of hydrogen sulphide.

6.

How is NO converted to NO2 for analysis in UV and Visible analyzers?(a) Treating sample gas with pressurized oxygen(b) Treating sample gas with ozone(c) Treating sample gas with oxygen at low pressure(d) Treating sample gas with water at high pressureI got this question in an international level competition.Question is from NO2 Analyser topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct option is (a) Treating sample GAS with pressurized oxygen

The best I can EXPLAIN: NO is converted into NO2 for ANALYSIS in UV and Visible ANALYZERS by treating sample gas with pressurized oxygen. This is because direct measurement of NO is not possible using UV and Visible analyzers.

7.

In which of the following ways can nitrogen be calibrated for dry gases?(a) By modifying the amplifier(b) By adjusting the needle valve(c) By modifying the detector(d) By changing chopper frequencyI have been asked this question during an interview.This key question is from Method Based on Ionization of Gases topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct choice is (b) By adjusting the needle VALVE

Explanation: NITROGEN can be CALIBRATED for dry gases by adjusting the needle valve. In this way, the ERROR will be compensated.

8.

In the Beckman instrument, the flow rate recommended when the sample enters directly for rapid response is which of the following?(a) 100-200 cc/min(b) 50-250 cc/min(c) 40-60 cc/min(d) 200-300 cc/minI got this question at a job interview.My enquiry is from Paramagnetic Oxygen Analyser topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right choice is (c) 40-60 cc/min

The BEST I can explain: In the Beckman instrument, the flow rate recommended when the SAMPLE enters directly for rapid RESPONSE is 40-60 cc/min. paramagnetic OXYGEN analyzer was first described by Pauling.

9.

Hydrocarbons need to be removed from the sample gas to avoid errors in Hartman and Braun oxygen analyzer.(a) True(b) FalseI had been asked this question during a job interview.My question is from Magnetic Wind Instruments in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct choice is (a) True

For explanation I WOULD say: HYDROCARBONS NEED to be removed from the sample gas to avoid errors. This is because they may react on the HEATED filaments and lead to a change in temperature and resistance.

10.

Which of the following bridges are used in thermal conductivity analyzers?(a) Wheatstone bridge(b) Kelvin’s bridge(c) Anderson’s bridge(d) Schering’s bridgeI got this question by my college professor while I was bunking the class.This interesting question is from Thermal Conductivity Analysers topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer» CORRECT OPTION is (a) Wheatstone bridge

For explanation: Wheatstone bridge is generally used in thermal CONDUCTIVITY analyzers. Themistors or hot-wires can be used as a filament.
11.

In which of the following conditions can thermal conductivity analyzer be used for multi-component mixtures?(a) When all the components have same thermal conductivities(b) When all the components have very different thermal conductivities(c) When all but one component have same thermal conductivities(d) When all components have very different thermal conductivities except two of them which are the sameI got this question in semester exam.I'm obligated to ask this question of Thermal Conductivity Analysers in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct CHOICE is (C) When all but one component have same thermal conductivities

The best explanation: When all but one component have the same thermal conductivities, thermal conductivity ANALYZER can be used for multi-component mixtures. The MIXTURE is then treated as a BINARY mixture.

12.

Which of the following is not the characteristic of thermistors used in thermal conductivity analyzers?(a) High negative temperature coefficient of resistance(b) Slow response(c) Extremely sensitive(d) Low costI have been asked this question during an interview for a job.I would like to ask this question from Thermal Conductivity Analysers in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

The CORRECT option is (b) SLOW response

Explanation: Thermistors have fast response.They are arranged as HEAT sensing ELEMENTS in the Wheatstone BRIDGE.

13.

Which of the following detectors are not used in analyzers based on gas density?(a) Hot wire(b) Platinum filament(c) Thermistor(d) ThermometerI have been asked this question in a job interview.My question is from Analysers Based on Gas Density topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right choice is (d) THERMOMETER

For explanation: Thermometer is not used as a detector in analyzers BASED on GAS density. Thermistors and HOT wires are GENERALLY used.

14.

How is NO converted to NO2 for analysis in Chemiluminescent analyser?(a) Treating sample gas with pressurized oxygen(b) Treating sample gas with ozone(c) Treating sample gas with oxygen at low pressure(d) Treating sample gas with water at high pressureI got this question in an international level competition.My enquiry is from NO2 Analyser topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right answer is (b) TREATING sample gas with ozone

Easy explanation: NO is converted to NO2 for analysis in CHEMILUMINESCENT analyser by treating sample gas with ozone.

NO + O3 —>NO2 + O2

15.

The instruments based on chemiluminescence maintain linearity in which of the following ranges?(a) 1ppb to 100ppb(b) 100ppb to 1000ppb(c) 1ppb to 1000ppb(d) 100ppb to 1000ppbThe question was asked in quiz.This question is from NO2 Analyser in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right OPTION is (C) 1ppb to 1000PPB

For explanation: The instruments based on chemiluminescence maintain linearity between 1ppb to 1000ppb. These are used to measure NO in exhaust GASES in vehicles.

16.

If the oxygen content is not greater than 21%, which of the following gases must be used to set the span point?(a) Oxygen(b) Dry air(c) Nitrogen(d) HydrogenThis question was posed to me by my school teacher while I was bunking the class.My doubt is from Paramagnetic Oxygen Analyser topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct ANSWER is (b) Dry air

Explanation: If the oxygen content is not greater than 21%, dry air is USED to set the SPAN point. This is done for calibration purposes.

17.

Which among the following elements has the highest thermal conductivity?(a) Nitrogen(b) Oxygen(c) Hydrogen(d) ChlorineThe question was asked by my school teacher while I was bunking the class.This question is from Thermal Conductivity Analysers topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right option is (c) HYDROGEN

The best I can EXPLAIN: Hydrogen has the HIGHEST thermal conductivity. Chlorine has the lowest thermal conductivity in the LIST.

18.

Changes in the composition of gas stream give rise to changes in thermal conductivity of the gas stream.(a) True(b) FalseThe question was posed to me in an interview for job.Question is taken from Thermal Conductivity Analysers in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct answer is (a) True

Easiest explanation: Changes in the COMPOSITION of gas stream give rise to changes in thermal CONDUCTIVITY of the gas stream. This principle is USED in thermal conductivity ANALYZERS.

19.

Polarographic cells type of electrochemical method uses which of the following concepts?(a) Cyclic reactions(b) Exothermic reactions(c) Reversible reactions(d) Redox reactionsThis question was posed to me in an internship interview.I'd like to ask this question from Electrochemical Methods for Oxygen Analysis topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct option is (d) Redox reactions

To explain I would say: Polarographic cells type of electrochemical METHOD USES redox reactions. It is USED to measure the partial pressure of the PERCENTAGE of oxygen.

20.

Nitrogen oxide cannot be directly analysed using UV and Visible analyzers due to which of the following reasons?(a) Less accuracy(b) Very low range(c) It leads to contamination of the sample(d) It is transparent in UV visible regionsI got this question in quiz.Enquiry is from NO2 Analyser topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct answer is (d) It is transparent in UV visible regions

For EXPLANATION I WOULD say: Nitrogen oxide cannot be directly analysed using UV and Visible ANALYZERS because it is transparent in UV visible regions. HENCE, it is converted into nitrogen dioxide and then analysed.

21.

Which of the following methods are mainly used in combustible gas detection equipment?(a) Lead acetate tape staining method(b) Solid state sensor(c) Gold film sensor(d) Electrochemical cellsThe question was asked in a job interview.I need to ask this question from H2S Analysers in section Industrial Gas Analysers of Analytical Instrumentation

Answer» RIGHT answer is (b) Solid state sensor

To EXPLAIN I would say: Solid state SENSORS are mainly used in combustible gas DETECTION equipment. It is a semiconductor sensor.
22.

Which kind of filter is used in improved version of infrared gas analyzer at the output end?(a) Low pass filter(b) High pass filter(c) Narrow bandpass filter(d) Wide bandpass filterThis question was addressed to me in an online interview.The above asked question is from Infrared Gas Analysers topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct option is (c) Narrow bandpass FILTER

To explain I would SAY: The filter is used in an improved VERSION of infrared gas ANALYZER at the output end is narrow bandpass filter. The bandpass characteristics are matched with the absorption SPECTRA of the gas of interest.

23.

In analyzers based on gas density, the unbalance of the bridge is linearly proportional to which of the following?(a) Temperature of sample detector(b) Pressure at sample detector(c) Pressure difference between both the detectors(d) Gas-density difference between both the detectorsThis question was posed to me by my college professor while I was bunking the class.My query is from Analysers Based on Gas Density topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct choice is (d) Gas-DENSITY difference between both the DETECTORS

To explain I would say: In analyzers based on gas density, the unbalance of the bridge is LINEARLY PROPORTIONAL to a gas-density difference between both the detectors. Instruments based on the PRINCIPLE of gas-density balance are commercially available.

24.

Which of the following measuring technique is utilized for measuring nitrogen?(a) By using a method based on gas density(b) By using paramagnetic analyzers(c) By using magnetic wind instruments(d) By using a photospectrometer where a gas sample is analysed and detected with a photocellThis question was addressed to me in unit test.Asked question is from Method Based on Ionization of Gases topic in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right choice is (d) By using a photospectrometer where a GAS sample is analysed and detected with a photocell

To EXPLAIN: By using a photospectrometer gas sample is analysed, for MEASURING nitrogen. It is then selectively FILTERED by using a photocell which provides an appropriate electric signal.

25.

To improve paramagnetic oxygen analyzer, the quartz suspension must be replaced with which of the following?(a) Platinum-iridium(b) Phosphor-bronze(c) Gold-palladium(d) AntimonyThe question was posed to me in a national level competition.My enquiry is from Paramagnetic Oxygen Analyser topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right option is (a) Platinum-iridium

For explanation: To IMPROVE PARAMAGNETIC oxygen analyzer, the quartz suspension MUST be REPLACED with Platinum-iridium. It is more ROBUST.

26.

The extremely high inertia of the thin membrane in the oxygen sensor allows it to rapidly attain equilibrium.(a) True(b) FalseThis question was posed to me during an interview.This question is from Magnetic Wind Instruments in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right option is (B) False

To explain I would say: The EXTREMELY low inertia of the thin membrane in the oxygen sensor allows it to ATTAIN EQUILIBRIUM. It reaches an equilibrium STATE in 0.25 to 0.5 seconds.

27.

Gases emit radiation under which of the following conditions which can be used for determination of gases?(a) When mixed with other gases(b) When kept at suitable temperature(c) With sufficient electrical excitation and at suitable pressures(d) When exposed to radiationThis question was posed to me in a job interview.The doubt is from Method Based on Ionization of Gases in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct option is (c) With SUFFICIENT electrical EXCITATION and at suitable pressures

For explanation: Gases emit radiation with sufficient electrical excitation and at suitable pressures. Gases emit radiation in different ways. It can be in the form of spark, ARC or GLOW DISCHARGE in different parts of the radiation spectrum.

28.

Polarographic cells are used to measure the partial pressure of gases only in static gas monitoring systems and not in continuous streams.(a) True(b) FalseThe question was asked in semester exam.My query is from Electrochemical Methods for Oxygen Analysis topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

The CORRECT option is (b) False

Explanation: Polarographic cells are not used to measure the partial PRESSURE of gases in static gas monitoring SYSTEMS alone. They can ALSO be used for measuring the partial pressure of gases in continuous samples and INJECTED samples.

29.

Which is the compound that forms the stain in Lead Acetate Tape Staining method used for the analysis of Hydrogen Sulphide and what is the colour of the stain?(a) Lead Sulphide, red colour(b) Lead Sulphide, brown colour(c) Hydrogen acetate, white colour(d) Hydrogen acetate, green colourThis question was posed to me in unit test.My question comes from H2S Analysers topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct ANSWER is (b) Lead Sulphide, BROWN colour

For EXPLANATION I would say: HYDROGEN Sulphide REACTS with Lead Acetate to form Lead Sulphide which is brown in colour. Hydrogen acetate in other words is acetic acid and is colourless.

30.

In Laser Opto-acoustic spectroscopy, the IR beam excites the molecules to higher states. In which of the following ways do the molecules return to the ground state?(a) Collisional de-excitation(b) Random de-excitation(c) By spontaneous emission(d) By stimulated emissionThis question was posed to me by my college professor while I was bunking the class.The origin of the question is NO2 Analyser in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct choice is (a) COLLISIONAL de-excitation

For explanation I WOULD say: The molecules return to the GROUND STATE by collisional de-excitation. This results in an increase in temperature.

31.

Nitrogen dioxide can be detected with a paramagnetic analyser.(a) True(b) FalseI had been asked this question in homework.This intriguing question comes from NO2 Analyser topic in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right choice is (a) True

To EXPLAIN: Nitrogen DIOXIDE can be detected with a paramagnetic analyser. This is because nitrogen has paramagnetic PROPERTY IE. it is attracted by magnetic fields.

32.

The diagram of Electrochemical cell used for the detection of hydrogen sulphide is given below. Identify the unmarked component.(a) Filter(b) Gas permeable membrane(c) Elastic layer(d) ElectrolyteI got this question in an interview for internship.My question is taken from H2S Analysers topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct option is (b) Gas PERMEABLE MEMBRANE

For explanation: The UNMARKED component is gas permeable membrane. Electrochemical CELL is an electrochemical gas diffusion sensor.

33.

Vibration excitation occurs only if the sample has hetero-atomic molecules.(a) True(b) FalseThis question was posed to me in class test.The above asked question is from Infrared Gas Analysers topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct answer is (a) True

Easiest explanation: VIBRATION excitation occurs only if the SAMPLE has hetero-atomic molecules. Hence, infrared analyzers are not used for GASES WHOSE molecules are FORMED by two identical atoms.

34.

When hot-wires are used as detectors amplification is not required.(a) True(b) FalseI got this question in class test.My doubt is from Analysers Based on Gas Density in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct answer is (b) False

The best EXPLANATION: When hot-wires are used as detectors amplification is REQUIRED. The AMPLIFIED SIGNAL is given to the recorder.

35.

Sample rates of nitrogen meter can be adjusted by doing which of the following?(a) By modifying the amplifier(b) With the help of the needle valve(c) By modifying the detector(d) By changing chopper frequencyThe question was posed to me by my college professor while I was bunking the class.This interesting question is from Method Based on Ionization of Gases in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right choice is (b) With the help of the needle valve

Explanation: SAMPLE rates of nitrogen meter can be ADJUSTED with the help of a needle valve. Modern nitrogen ANALYZERS have digital DISPLAYS.

36.

Nitrogen meters are calibrated for which of the following to measure nitrogen?(a) Mixture of nitrogen and air(b) Water saturated mixture of nitrogen and oxygen(c) Mixture of nitrogen and hydrogen(d) NitrogenThe question was posed to me during a job interview.The origin of the question is Method Based on Ionization of Gases in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer» CORRECT choice is (B) Water saturated mixture of nitrogen and oxygen

Easiest explanation: Nitrogen METERS are calibrated for water saturated mixture of nitrogen and oxygen. Hence, this calibration cannot be used for DRY GASES.
37.

The light output from the discharge tube is interrupted by means of a rotating slotted disc.(a) True(b) FalseI got this question in an online quiz.My doubt stems from Method Based on Ionization of Gases in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct answer is (a) True

Best explanation: The LIGHT OUTPUT from the discharge TUBE is interrupted by means of a ROTATING slotted disc. Hence, a chopped output is obtained.

38.

Which of the following gases have paramagnetic property (ability to get attracted to a magnetic field)?(a) Nitric oxide(b) Hydrogen(c) Helium(d) NitrogenI have been asked this question in an online interview.Asked question is from Paramagnetic Oxygen Analyser in division Industrial Gas Analysers of Analytical Instrumentation

Answer» RIGHT answer is (a) Nitric oxide

For explanation: Nitric oxide has PARAMAGNETIC property. Nitrogen dioxide and OXYGEN are other gases that have paramagnetic property.
39.

The force produced during the operation of Paramagnetic oxygen analyzer is proportional to which of the following?(a) Magnetic susceptibility of sphere(b) Magnetic susceptibility of surrounding gas(c) Difference between magnetic susceptibility of sphere and magnetic field strength(d) Difference between magnetic susceptibility of sphere and that of surrounding gasThe question was asked in class test.Origin of the question is Paramagnetic Oxygen Analyser topic in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct option is (d) Difference between magnetic SUSCEPTIBILITY of SPHERE and that of surrounding gas

The explanation is: The PARAMAGNETIC property of oxygen is used in this method for the analysis of oxygen. The force produced during operation of Paramagnetic oxygen analyzer is PROPORTIONAL to the difference between magnetic susceptibility of sphere and that of surrounding gas.

40.

In the Beckman instrument, the flow rate recommended when the sample enters through a porous diffusion disc is which of the following?(a) 100-200 cc/min(b) 50-250 cc/min(c) 40-60 cc/min(d) 200-300 cc/minI got this question in an interview for internship.The above asked question is from Paramagnetic Oxygen Analyser in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct choice is (B) 50-250 cc/min

To explain: In the Beckman instrument, the flow rate recommended when the sample enters through a porous diffusion DISC is 50-250 cc/min. PARAMAGNETIC oxygen analyzer was FIRST DESCRIBED by Pauling.

41.

Which of the following does not occur in Hartman and Braun oxygen analyzer based on the magnetic wind?(a) Oxygen is attracted to the magnetic field(b) Oxygen becomes less paramagnetic due to rise in temperature(c) Resistance change in a resistance connected to the Wheatstone bridge occur(d) Wheatstone Bridge becomes balanced due to temperature changeThe question was asked in semester exam.Origin of the question is Magnetic Wind Instruments topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct choice is (d) WHEATSTONE BRIDGE becomes BALANCED due to TEMPERATURE change

Explanation: In Hartman and Braun oxygen analyzer, the magnetic wind COOLS the heated winding. As a result, the bridge becomes unbalanced.

42.

In the gold film sensor, the change in resistance is proportional to the concentration of H2S.(a) True(b) FalseThis question was addressed to me in class test.The above asked question is from H2S Analysers in division Industrial Gas Analysers of Analytical Instrumentation

Answer» RIGHT choice is (a) True

The BEST I can explain: Gold film sensor works on the principle of ABSORPTION. The change in RESISTANCE is PROPORTIONAL to the concentration of H2S.
43.

In calibration of CO2 analyser, in order to establish zero calibration which of the following is used as sample?(a) Oxygen(b) Hydrogen(c) Inert gases(d) NitrogenI had been asked this question in an online interview.My doubt is from Infrared Gas Analysers topic in portion Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct choice is (c) INERT gases

The EXPLANATION: In calibration of CO2 analyser, in order to establish ZERO calibration inert gases are used as sample. Generally, room AIR is also used.

44.

Which of the following is true about the two charged vanes that produce electric field in Paramagnetic oxygen analyzer?(a) Both the vanes should be at higher potential than the test body(b) Both the vanes should have the same potential as the test body(c) One vane should be at higher potential than the test body and the other vane should be at lower potential(d) Both the vanes should be at lower potential than the test bodyThe question was asked during a job interview.My question is taken from Paramagnetic Oxygen Analyser topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct answer is (c) One VANE should be at HIGHER potential than the test body and the other vane should be at lower potential

To EXPLAIN I would say: In Paramagnetic oxygen analyzer, the two CHARGED vanes should be such that one vane should be at higher potential than the test body and the other vane should be at lower potential. This electric FIELD produces a force.

45.

In OXYMAT-M oxygen analyser which of the following is usually used as reference gas?(a) Hydrogen(b) Nitrogen(c) Helium(d) Ambient airI have been asked this question in an internship interview.Asked question is from Magnetic Wind Instruments in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct CHOICE is (d) Ambient air

The explanation: In OXYMAT-M oxygen analyser, ambient air is usually used as a REFERENCE gas. Its measuring PROCEDURE is based on PRESSURE difference which DEVELOPS between two gases having different oxygen concentration in the magnetic field.

46.

During analysis of NO2 using Chemiluminescent analyser, why is NO2 not made to react with ozone directly?(a) Less accuracy(b) It is a slow process(c) It leads to contamination of the sample(d) It does not produce luminescenceThis question was addressed to me in an interview for job.My doubt is from NO2 Analyser topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right option is (b) It is a slow process

The BEST I can explain: During analysis of NO2 using Chemiluminescent analyser, NO2 is not made to REACT with ozone directly because it is a slow process. Hence, it is CONVERTED to NO using CATALYTIC reactions or converters.

47.

How can absorption be enhanced while determining NO concentration using CO laser?(a) By converting NO into NO2(b) By placing NO in a magnetic field(c) By using proper monochromators(d) By using choppersI have been asked this question at a job interview.This interesting question is from NO2 Analyser in section Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right answer is (b) By PLACING NO in a magnetic field

Easiest EXPLANATION: CO LASER emits RADIATION that can absorb NO. Absorption is enhanced by placing NO in a magnetic field.

48.

Polarographic cells are not sensitive to which of the following gases?(a) Carbon monoxide(b) Carbon dioxide(c) Nitrous oxide(d) OxygenThe question was posed to me in an interview for internship.Asked question is from Electrochemical Methods for Oxygen Analysis in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Correct choice is (a) Carbon monoxide

Easy explanation: Polarographic CELLS are not SENSITIVE to carbon monoxide. They SHOW 0.1% error while measuring oxygen.

49.

Which of the following is not an advantage of solid state sensor used for analysis of H2S?(a) No sampling system(b) It can be used in conditions involving vibration(c) It can be used in corrosive atmosphere(d) No interference from background gasesThis question was addressed to me in exam.The origin of the question is H2S Analysers topic in division Industrial Gas Analysers of Analytical Instrumentation

Answer»

Right answer is (d) No INTERFERENCE from background gases

Explanation: SOLID state sensors are affected by interference from background gases. Some of them are HYDROGEN, isopropanol and ethyl and methyl mercaptan.

50.

Which of the following is the chopping frequency used for industrial analyzers in the simple infrared analyser for gas analysis?(a) 2-10 Hz(b) 2-50 Hz(c) 2-100 Hz(d) 2-150 HzThis question was posed to me during an interview for a job.Question is taken from Infrared Gas Analysers topic in chapter Industrial Gas Analysers of Analytical Instrumentation

Answer»

The correct CHOICE is (a) 2-10 HZ

Explanation: The chopping FREQUENCY used for industrial analysers in the simple infrared analyzer for gas ANALYSIS is 2-10 Hz. The chopper is motor driven.