How Do You Calibrate A Fieldbus Instrument?

Answer»
  • This is really no different to any other smart transmitter, except that now, you don’t have to worry about the 4-to-20mA analogue signal! You just need to read out the measurement over the bus with a suitable HOST device.
  • You need a source of the process variable: PRESSURE, temperature (or ohms or millivolts), pH, or whatever. Typically, when you connect a known process value to the sensor, you can send a special MESSAGE to the device, INCLUDING the value you read, and what it should be. The field device will then adjust a stored parameter to make it so. Two values, near the ends of the instrument’s working range, are usually enough, though some devices may USE a more complex multi-point calibration in the factory to adjust linearity corrections, or even a temperature cycle to adjust temperature correction.
  • Automatic calibrators will step through this process with only minimal user involvement.